Hello
I'm trying to understand the accuracy of a PAT tester which states the resolution, % and digit tolerances of individual test types. My calculations are below for a hypothetical test and I would be very grateful if anyone could tell me if I'm doing right. Resolution = 0.01 (2dp) Accuracy = ±5% ±2 digits Test Reading = 15.50 Step 1: Determine the % tolerance 15.50 * 0.05 = 0.775 Step 2: Round to 2dp = 0.78 Step 3: Add the digit accuracy to least significant figure = 0.78 + 0.02 =0.8 Therefore for a reading of 15.50 ± 0.8 (5.16% of 15.50) Upper limit = 16.30 Lower limit = 14.70 Thanks for your time :-) Lorna |
|
|
|
|
|
|
|
|
Lorna,
Afraid I don't understand the Jargon in your Post - a bit before my time - but there is a sad lack of learning in the matter of estimating errors. Are you aware thjqt the first - and possibly the greatest - error lies in reading the test meter? All meters should be indicating as near as possible to their f.s.d. ? Ken Green |
|
|
|
|
|
|
|
|
Lorna:
That working looks right to me. Ken: By the sound of it, that's a digital and not an analogue meter. So the concept of FSD and meter reading errors wouldn't really apply. ------------------------- S P Barker BSc PhD MIET |
|
|
|
|
|
|
|
|
Add AND SUBTRACT the 2%. because tolerances and errors are usually +/-..
Mind you, it could be worth carefully re-reading the spec because with a digital meter you should not have a reading error - it is reading 15.5 v or it is not. As ectophile says....... There is no parallax error as with an analogue meter ------------------------- Alan n.b. This message has a limited shelf life. |
|
|
|
|
|
|
|
|
Accuracy = ±5% ±2 digits
That might be Accuracy = ±5% or ±2 digits - which ever is larger (rather than both). The values you have been given are the nominal values given by the manufacturer and don't necessarily reflect the machine you have - do you have, or can you get, the calibration certificate? |
|
|
|
|
|
|
|
|
Hi to All,
I seem to have sparked quite a discussion - and who can complain about that? It was mainly that Lorna's question reminded me of the budding amateur who settled on my visitors' stool, examined my Bench and then enquired whether he should purchase a digital or an analogue meter? My attempts to explain obviously were confusing him and I deemed it best to let him grow a little. Ken |
|
|
|
|
|